Abstract
Background
Medical doctors who perform C-arm fluoroscopy-guided procedures are exposed to X-ray radiation. Therefore, radiation-protective shields are recommended to protect these doctors from radiation. For the past several years, these protective shields have sometimes been used without regular inspection. The aim of this study was to investigate the degree of damage to radiation-protective shields in the operating room
Methods
This study investigated 98 radiation-protective shields in the operation rooms of Konkuk University Medical Center and Jeju National University Hospital. We examined whether these shields were damaged or not with the unaided eye and by fluoroscopy
Results
There were seventy-one aprons and twenty-seven thyroid protectors in the two university hospitals. Fourteen aprons (19.7%) were damaged, whereas no thyroid protectors (0%) were. Of the twenty-six aprons, which have been used since 2005, eleven (42.3%) were damaged. Of the ten aprons, which have been used since 2008, none (0%) was damaged. Of the twenty-three aprons that have been used since 2009, two (8.7%) of them were damaged. Of the eight aprons used since 2010, one (12.3%) was damaged. Of the four aprons used since 2011, none (0%) of them were damaged. The most common site of damage to the radiation-protective shields was at the waist (51%) of the aprons
Conclusions
As a result, aprons that have been used for a long period of time can have a higher risk of damage. Radiation-protective shields should be inspected regularly and exchanged for new products for the safety of medical workers